Re: Repeatability, Accuracy or Resolution
Here's another consideration, for what it's worth.
A Wixey measures the orientation of a surface relative to the gravitational vector. I have no reason to doubt that it does so within the claimed accuracy of +/- 0.1 degrees.
Now when you go and measure the orientation of a second surface, that orientation is also measured with respect to the gravitational vector. If what you really want to know is the orientation of Surface B relative to Surface A, all you have to do is subtract the two gravity-based measurements, just like in Dusty's example.
Except: both readings have a possible error of 0.1 degrees. Those errors may offset one another, or they may combine to double the relative-angle error. Bottom line, the guaranteed accuracy of the relative-angle measurement is now +/- 0.2 degrees.
A Wixey measures the orientation of a surface relative to the gravitational vector. I have no reason to doubt that it does so within the claimed accuracy of +/- 0.1 degrees.
Now when you go and measure the orientation of a second surface, that orientation is also measured with respect to the gravitational vector. If what you really want to know is the orientation of Surface B relative to Surface A, all you have to do is subtract the two gravity-based measurements, just like in Dusty's example.
Except: both readings have a possible error of 0.1 degrees. Those errors may offset one another, or they may combine to double the relative-angle error. Bottom line, the guaranteed accuracy of the relative-angle measurement is now +/- 0.2 degrees.
Re: Repeatability, Accuracy or Resolution
Tell me if this is the best way.
To get a bevel angle, I zero the Wixey on the blade then move it to the (already closely angled) table. Then adjust the table to the angle needed.
I wish magnets worked on aluminum.
To get a bevel angle, I zero the Wixey on the blade then move it to the (already closely angled) table. Then adjust the table to the angle needed.
I wish magnets worked on aluminum.
- JPG
- Platinum Member
- Posts: 35598
- Joined: Wed Dec 10, 2008 7:42 pm
- Location: Lexington, Ky (TAMECAT territory)
Re: Repeatability, Accuracy or Resolution
Stiction in the measuring device(the wixey 'bearings').jsburger wrote:I think that is what I have been trying to say. Dusty seems to disagree for some reason.JPG wrote:Moving a measurement device about an irregular surface and getting differing readings has nothing to do with accuracy or repeatability. More than likely the irregularities will exceed the influences of technique.
When using the wixey(or similar), stiction has to be allowed for and is likely the determining factor of all it's accuracy deviations.
I don't know why stiction is a factor. Stiction is the FORCE required to to overcome static cohesion and slide two stationary objects in contact with each other on a parallel surface. How does that translate to reading differences?
╔═══╗
╟JPG ╢
╚═══╝
Goldie(Bought New SN 377425)/4" jointer/6" beltsander/12" planer/stripsander/bandsaw/powerstation /Scroll saw/Jig saw /Craftsman 10" ras/Craftsman 6" thicknessplaner/ Dayton10"tablesaw(restoredfromneighborstrashpile)/ Mark VII restoration in 'progress'/ 10E[/size](SN E3779) restoration in progress, a 510 on the back burner and a growing pile of items to be eventually returned to useful life. - aka Red Grange
╟JPG ╢
╚═══╝
Goldie(Bought New SN 377425)/4" jointer/6" beltsander/12" planer/stripsander/bandsaw/powerstation /Scroll saw/Jig saw /Craftsman 10" ras/Craftsman 6" thicknessplaner/ Dayton10"tablesaw(restoredfromneighborstrashpile)/ Mark VII restoration in 'progress'/ 10E[/size](SN E3779) restoration in progress, a 510 on the back burner and a growing pile of items to be eventually returned to useful life. - aka Red Grange
- BuckeyeDennis
- Platinum Member
- Posts: 3812
- Joined: Tue Jul 24, 2012 10:03 pm
- Location: Central Ohio
Re: Repeatability, Accuracy or Resolution
I don't think a Wixey is a pendulum-type device. Probably based on a 2- or 3-axis MEMS accelerometer, with digital calibration curves. So no stiction.JPG wrote:Stiction in the measuring device(the wixey 'bearings').jsburger wrote:I think that is what I have been trying to say. Dusty seems to disagree for some reason.JPG wrote:Moving a measurement device about an irregular surface and getting differing readings has nothing to do with accuracy or repeatability. More than likely the irregularities will exceed the influences of technique.
When using the wixey(or similar), stiction has to be allowed for and is likely the determining factor of all it's accuracy deviations.
I don't know why stiction is a factor. Stiction is the FORCE required to to overcome static cohesion and slide two stationary objects in contact with each other on a parallel surface. How does that translate to reading differences?
- JPG
- Platinum Member
- Posts: 35598
- Joined: Wed Dec 10, 2008 7:42 pm
- Location: Lexington, Ky (TAMECAT territory)
Re: Repeatability, Accuracy or Resolution
Howcum it acts like it?BuckeyeDennis wrote: . . .
I don't think a Wixey is a pendulum-type device. Probably based on a 2- or 3-axis MEMS accelerometer, with digital calibration curves. So no stiction.
Well my igage one does.
╔═══╗
╟JPG ╢
╚═══╝
Goldie(Bought New SN 377425)/4" jointer/6" beltsander/12" planer/stripsander/bandsaw/powerstation /Scroll saw/Jig saw /Craftsman 10" ras/Craftsman 6" thicknessplaner/ Dayton10"tablesaw(restoredfromneighborstrashpile)/ Mark VII restoration in 'progress'/ 10E[/size](SN E3779) restoration in progress, a 510 on the back burner and a growing pile of items to be eventually returned to useful life. - aka Red Grange
╟JPG ╢
╚═══╝
Goldie(Bought New SN 377425)/4" jointer/6" beltsander/12" planer/stripsander/bandsaw/powerstation /Scroll saw/Jig saw /Craftsman 10" ras/Craftsman 6" thicknessplaner/ Dayton10"tablesaw(restoredfromneighborstrashpile)/ Mark VII restoration in 'progress'/ 10E[/size](SN E3779) restoration in progress, a 510 on the back burner and a growing pile of items to be eventually returned to useful life. - aka Red Grange
- dusty
- Platinum Member
- Posts: 21530
- Joined: Wed Nov 22, 2006 6:52 am
- Location: Tucson (Wildcat Country), Arizona
Re: Repeatability, Accuracy or Resolution
I don't think so. Read "How Does It Read". The specifications might be enlightening too.JPG wrote:Stiction in the measuring device(the wixey 'bearings').jsburger wrote:I think that is what I have been trying to say. Dusty seems to disagree for some reason.JPG wrote:Moving a measurement device about an irregular surface and getting differing readings has nothing to do with accuracy or repeatability. More than likely the irregularities will exceed the influences of technique.
When using the wixey(or similar), stiction has to be allowed for and is likely the determining factor of all it's accuracy deviations.
I don't know why stiction is a factor. Stiction is the FORCE required to to overcome static cohesion and slide two stationary objects in contact with each other on a parallel surface. How does that translate to reading differences?
http://www.wixey.com/anglegauge/faq/index.html#03
"Making Sawdust Safely"
Dusty
Sent from my Dell XPS using Firefox.
Dusty
Sent from my Dell XPS using Firefox.
- BuckeyeDennis
- Platinum Member
- Posts: 3812
- Joined: Tue Jul 24, 2012 10:03 pm
- Location: Central Ohio
Re: Repeatability, Accuracy or Resolution
Because my hunch was wrong!JPG wrote:Howcum it acts like it?BuckeyeDennis wrote: . . .
I don't think a Wixey is a pendulum-type device. Probably based on a 2- or 3-axis MEMS accelerometer, with digital calibration curves. So no stiction.![]()
Well my igage one does.
From Dusty's "How Does It Read" link, a Wixey does appears to be a pendulum-type device, with a non-contacting angle sensor. Except they call it a "rotating counterweight".
Back around 1980, I used a fairly high-end oil-damped pendulum sensor on a robotics project. Accuracy on that was only about half a degree, IIRC. But the sensor on that device was a potentiometer, which was probably the main source of stiction. I later replaced it with good accelerometers, which were quite expensive back then.
- dusty
- Platinum Member
- Posts: 21530
- Joined: Wed Nov 22, 2006 6:52 am
- Location: Tucson (Wildcat Country), Arizona
Re: Repeatability, Accuracy or Resolution
What you are doing is correct and will give you the proper tilt angle but there is one thing to consider.Gene Howe wrote:Tell me if this is the best way.
To get a bevel angle, I zero the Wixey on the blade then move it to the (already closely angled) table. Then adjust the table to the angle needed.
I wish magnets worked on aluminum.
You must maintain an angular relationship between the position of the Wixey when attached to the blade and the Wixey when attached to the tilted table.
To do what you are saying, I use my Incra 120 Miter Gauge. The protractor is magnetic. You can set the Wixey on the protractor and it will adhere. With the miter gauge set at 90 degrees and the Wixey positioned against the vertical surface of the protractor, you have established that relationship.
Now when you tilt the table, the Wixey will give you an accurate tilt angle.
What happens if you do not maintain that angular relationship? Simply unlock the protractor and rotate the protractor to something other than 90 degrees. You will see the Wixey "tilt angle" change but the tilt angle of the main table has not. This is, in my opinion, a technique error.
"Making Sawdust Safely"
Dusty
Sent from my Dell XPS using Firefox.
Dusty
Sent from my Dell XPS using Firefox.
Re: Repeatability, Accuracy or Resolution
I have a Wixey-like Beall Tilt Box. I use it for angles other than 90/45.
For 90 degrees, I now prefer a small 4 inch engineer's square. For 45 degrees, I use a small 45 degree square like this one.

For 90 degrees, I now prefer a small 4 inch engineer's square. For 45 degrees, I use a small 45 degree square like this one.

Gale's Law: The bigger the woodworking project, the less the mistakes show in any photo taken far enough away to show the entire project!
Re: Repeatability, Accuracy or Resolution
Thanks Dusty. That makes sense. I never thought of using the miter gauge too.